Re: Scanning encoder debounce on new scope
OK, if you just want to see squiggly lines on the screen :-)
But you're not going to mess it up. It's part of verifying that it actually works as it should.
Compensating the probe, or at least veryfying its compensation, should ideally be done whenever you move a probe from one channel to another - which is another great reason to use the color coded rings on the probes so that the same probe gets used on the same channel each time.
Without a properly compensated probe your scope will not display a "true" representation of the waveform and things like automatic risetime and peak-to-peak voltage measurements might be incorect.
I'm sure there's a section, close to the beginning, in the manual on how to perform the probe compensation.